LEAK RATE 4.0

I AM GETTING A LEAK RATE OF 4.0. I REBUILT THE STEAM MODULE, ALSO THE EXHAUST MODULE. NEW GASKET, CHECK VALVES ON THE HEAT EXCHANGER. NEW HEAT EXCHANGER. COULD I BE GETTING THE LEAK THROUGH THE FLOAT SWITCH INSIDE THE CHAMBER?


   RE: Sterilizers/Autoclaves Report Post
Sun Aug 20 2023Reply from Reliable Sterilizer Services
avatar placemark

Rebuilt the S1 valve, also inspect the check valve .


Report Post
Sun Aug 20 2023Reply from Reliable Sterilizer Services
avatar placemark

That's a big leak, pressurize the chamber and use a mirror to inspect for any steam leaks.


Report Post
Sun Aug 20 2023Reply from Reliable Sterilizer Services
avatar placemark

What's the make and model?


Report Post
Mon Aug 21 2023Reply from UNM
avatar placemark

Steris v116.


Report Post
Mon Aug 21 2023Reply from Sal Reyes
avatar placemark

Check if the door is correctly alligned... I had a similar issue with a getinge 433, the problem was that the door on the right side came out a little more than the left side, when I calibrate the guidelines with a sparkplug calibrator, the leak stop. Also, check the gasket itself, if seals good. Perhalps you already did it, but check the weld and pipelines.


Report Post
Thu Sep 07 2023Reply from UNM
avatar placemark

SAL REYES, I'LL CHECK THE DOOR, SEEMS LIKE THAT'S THE ONLY THING I HAVE LEFT TO TRY!


Report Post
Tue Oct 31 2023Reply from lorde
avatar placemark

A leak rate of 4.0 in an autoclave, like the Tuttnauer EZ11 Plus, is a significant concern, as it can indicate issues with the sterilization process. You've already taken several steps to address potential sources of the leak, including rebuilding the steam and exhaust modules, replacing gaskets, and checking valves on the heat exchanger.


buckshot roulette
Report Post
Mon Nov 20 2023Reply from Chris-CBET
Chris-CBET

Best thing to do is add hand valves to the system, that way you can isolate.  


Report Post
Mon Nov 20 2023Reply from UNM
avatar placemark

Thanks for the response. I found out the seats on the steam manifold, the o-ring was worn out. Leaking by.


Report Post