LEAK RATE 4.0

I AM GETTING A LEAK RATE OF 4.0. I REBUILT THE STEAM MODULE, ALSO THE EXHAUST MODULE. NEW GASKET, CHECK VALVES ON THE HEAT EXCHANGER. NEW HEAT EXCHANGER. COULD I BE GETTING THE LEAK THROUGH THE FLOAT SWITCH INSIDE THE CHAMBER?


   RE: Sterilizers/Autoclaves Report Post
Sun Aug 20 2023Reply from Reliable Sterilizer Services
avatar placemark

Rebuilt the S1 valve, also inspect the check valve .


Report Post
Sun Aug 20 2023Reply from Reliable Sterilizer Services
avatar placemark

That's a big leak, pressurize the chamber and use a mirror to inspect for any steam leaks.


Report Post
Sun Aug 20 2023Reply from Reliable Sterilizer Services
avatar placemark

What's the make and model?


Report Post
Mon Aug 21 2023Reply from UNM
avatar placemark

Steris v116.


Report Post
Mon Aug 21 2023Reply from Sal Reyes
avatar placemark

Check if the door is correctly alligned... I had a similar issue with a getinge 433, the problem was that the door on the right side came out a little more than the left side, when I calibrate the guidelines with a sparkplug calibrator, the leak stop. Also, check the gasket itself, if seals good. Perhalps you already did it, but check the weld and pipelines.


Report Post
Thu Sep 07 2023Reply from UNM
avatar placemark

SAL REYES, I'LL CHECK THE DOOR, SEEMS LIKE THAT'S THE ONLY THING I HAVE LEFT TO TRY!