Hi All, Can anyone shed a light on what causes the above error. I have searched everywhere and can find no mention of it?
Equipment: Covidien - Nellcor PM100N
Hi All, Can anyone shed a light on what causes the above error. I have searched everywhere and can find no mention of it?
Equipment: Covidien - Nellcor PM100N
If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.
Just be sure to tell them you were referred by MedWrench
I've had three PM100N's with this EEE901 error code in the log. It can be caused by either the Multi-Functional Respiratory (SpO2) board (MFR board), or by a faulty SPO2 sensor. Unfortunately, the end user didn't return the SPO2 sensors so I couldn't verify that the sensor was the cause, therefore, I left the PM 100N's soak test over a few days to see if the fault reoccurred. It only reoccurred on one device and replacing the MFR board cleared this. The other two were put back into service with new SPO2 sensors.
Hope this helps.
Kind Regards
JB