Wed Aug 16 2023

[7963] no communication with ergometer/treadmile

[7963] no communiction with  ergometer /treadmile

Equipment: Cardiac Science - Quinton TM55 & TM65 Treadmill


   RE: Treadmill Report Post
Wed Aug 16 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

Agiliti

 

Just be sure to tell them you were referred by MedWrench


Report Post
Thu Aug 17 2023Reply from Amscopolaris
Amscopolaris

 may be a problem with theUSB communication cable. or Check com port settings


Report Post
Sat Aug 19 2023Reply from Denova Medical
Denova Medical

First check or replace the interface cable between the treadmill and the controller. That cable has the problem of getting under the cover of the treadmill and getting cut,


Photios Peter Dalamagas Denova Medical Inc (704)-233-7574