Amps total of the equipment

Dornoch ultra you’re saying a 15 amp circuit but what is the total amperage of the circuit going to be?

Equipment: Zimmer - Dornoch Ultra Evac


   RE: Waste Management System Report Post
Thu Aug 17 2023Reply from Chris S
Chris S

Electrical The unit's electrical supply must be within 36" (92cm) of the unit location. A ground fault receptacle rated at 120 VAC, single phase at 60 Hz and 15 amps is required. The Ultra Evac Unit draws a peak current of 8.8 amps with a normal operating current draw of 0.2 to 6.0 amps.