Thu Aug 24 2023
Front panel assembly swap
So I swapped the front panel assembly and now the new piece does not turn on. I checked with the old part and no power there either. The board gets power but the board reset button does not work. Any insight on what steps need to be taken if any when swapping that part.
   RE: Centrifuge Report Post
Thu Aug 24 2023Reply from Chris-CBET
do you have a multimeter?
Fri Aug 25 2023Reply from TRS
First of all, double check the ribbon cable connection.  Secondly, if the old touch pad and the new touch pad are failing in the exact same manner, the front panel is most likely not the issue.  There can be an issue with the power supply or the main pcb.  Output voltage to the main pcb needs to be checked firstly...you should be able to measure 23.91 to 24.91 vdc.  If that voltage is present, the main pcb is most likely the culprit.