IU22 keep turning off by itself

IU22 ultrasound keeps turning off by itself after its turned on and it display an error just before it goes off that Supported video recorder is not selected

Equipment: Philips - iU22


   RE: Ultrasound Systems Report Post
Mon Sep 11 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

Encore Technical Services

Integris Equipment

MERCO Biomedical

 

Just be sure to tell them you were referred by MedWrench


Report Post
Tue Sep 12 2023Reply from FENC
avatar placemark

The DVR message can be ignored. However, this sounds like the system is getting the power down signal from either the USB interface behind the DVD drive or the power and signal board on the wall of the backend. 


I have seen where fluid infiltration and corrosion from Cavi-wipes being set on the side under the monitor articulation arm which causes the fluid to leak down into that area. 


Report Post
Sun Sep 24 2023Reply from Denova Medical
Denova Medical

You need to eliminate a failing power supply first. Software controlled equipment with many components controlled by the software need to boot up each component in a sequence. As each component boots up, more load is applied to the power supply and to the various power supply voltage rails. A weakened power supply could fail during the boot up sequence, in your case, during the video recorder boot up.


Photios Peter Dalamagas Denova Medical Inc (704)-233-7574