Error i18

Please help troubleshooting thia error

Equipment: GE HealthCare - Senographe Essential


   RE: Mammography Report Post
Mon Oct 30 2023Reply from MapleTreeBiomed
MapleTreeBiomed

Hi RHCB, 

The manual says I18 is "Collimator/Console locked" and is due to "The X-ray Console and the diaphragm control are locked due to service access" and recommends "Wait until end of service access. If no service in progress power OFF/ON from the X-ray Console." After that I'd call GE.

Good Luck!


Andrew Whyte, CBET, Maple Tree Biomedical LLC