Door unsealed alarm

Hi

Need help on Sm Century 400 with door unsealed alarm. Exhaust and steam manifold have been changed twice. CK1/2 and 14 also have been changed. PRV, door pressure switch also change and calibrated. After starting DART test not even 5 minutes after the seal got activated the error shows up door unsealed alarm. Im struggling to get it resolve and run out of option hence reaching out here for help. 

Equipment: Amsco 400 small


   RE: Sterilizers/Autoclaves Report Post
Wed Apr 03 2024Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

MERCO Biomedical

 

Just be sure to tell them you were referred by MedWrench     


Report Post
Sun Jun 09 2024Reply from Chris-CBET
Chris-CBET

the door sensors are magnetic sensors. you'll see one on the top left of the door, the other on the bottom right near the foot switch. if the light goes off, it's lost contact with the piece of metal the extends from the door. a simple adjustment of the sensor can solve your issue after you've ruled out door seal leak, steam quality


Report Post
Thu Jun 13 2024Reply from Cbet_Matt
Cbet_Matt

The century and 400 units I have the door closed. Switch is a proximity sensor along with the door activation foot switch. You might want to check the exhaust manifold and make sure that all the ports are clear coming out and going to the pressure switch. Is the voltage to S 35 solid?