I have this error ,when the KV reaches more than 64 ,then this error appear.
E 1066 can fault line active.
Any help please
Any help please
No dear..but the pump was not working.. i remove it.. the xray will.be normal in the auto on air...but when i put paitent and the KV exceeds 64.then this merror appear E1066
This means generator has an internal arc issue or the control is failing. Feel free to reach out directly to us for more assistance.
Whenever experiencing 1066 it is likely tube is actually cracked internally. Re-Calibration or calibration adjustment will not solve this problem. Please contact us direct if you need a repair for it.
Dear could you please send me the instruction for tune conditioning process.how it can be done.
Please to email me on :