E 1066 can fault line active. ziehm solo

I have this error ,when the KV reaches more than 64 ,then this error appear.

E 1066 can fault line active.


Any help please


   RE: Mobile C-Arm Report Post
Thu May 30 2024Reply from The Ziehm Guy
avatar placemark

Are there any other errors in the error log


Report Post
Thu May 30 2024Reply from Eng Hassan
avatar placemark

No dear..but the pump was not working.. i remove it.. the xray will.be normal in the auto on air...but when i put paitent and the KV  exceeds 64.then this merror appear E1066


Report Post
Thu May 30 2024Reply from Neomed Imaging - Service Dept
Neomed Imaging - Service Dept

This means generator has an internal arc issue or the control is failing. Feel free to reach out directly to us for more assistance.


NEOMED IMAGING - SERVICE - Ziehm C-Arm Parts, Repairs, Sales - Phone: 951-742-7020 - www.neomed-imaging.com
Report Post
Thu May 30 2024Reply from Tomelgy
avatar placemark

E1066 is CAN error, could be tube arcing.

Try to run the tune conditioning process


Report Post
Thu May 30 2024Reply from Neomed Imaging - Service Dept
Neomed Imaging - Service Dept

Whenever experiencing 1066 it is likely tube is actually cracked internally. Re-Calibration or calibration adjustment will not solve this problem. Please contact us direct if you need a repair for it.


NEOMED IMAGING - SERVICE - Ziehm C-Arm Parts, Repairs, Sales - Phone: 951-742-7020 - www.neomed-imaging.com
Report Post
Fri May 31 2024Reply from Eng Hassan
avatar placemark

Dear could you please send me the instruction for tune conditioning process.how it can be done.

Please to email me on :

Hassantahini@hotmail.com


Report Post