Sat Jun 01 2024
Support
Please I have a faulty Arcadis Varic VC 10A with error message 4131and 4121.
And the 2 folded monitors on arrival refused to open
Equipment: Service manual or technical support
Mon Jun 03 2024Reply from Tomelgy
MSY 4121 is the Image system failed to start properly. Check connection to PC rear D66 (CIPP) to D1 CPU X8, SG cable/connector.
MSY 4131 is a general time-out message.
When you say both monitors will not swing-out, do you mean physically or do you mean both monitors are not showing any text.
iF no text, check the PC has started, do you see any boot-up messages etc.
Tom
Tue Jun 04 2024Reply from prof101
Yes I mean the 2 monitors can not open physically and both error 4121 and 4131 are showing on the c-arm, not the main screen. Please can you add me  so I can send you pictures or videos.   +2348082255114
Tue Jun 04 2024Reply from Tomelgy
  • You need to be able to see both monitor screens inorder to attempt identify your errors 4121 & 4131.
The monitors not opening is unusual, you should be able to dismount the monitors to view screens
Siemens have a website where documents can be downloaded.
Register then search Arcadis
Was hoping you will understand me better if you see pictures
Wed Jun 05 2024Reply from Tomelgy
Yes, understood. But outside Mediwrench is not permitted
Wed Jun 05 2024Reply from prof101
If I can send images and videos here, I wouldn't ask for number.
I have unmounted the screen and I don't think the system software is loading but not sure if that's also the cause of error 4131 and 4121