Asking about two error codes appear on the control panel 6531 & 6910

Hi Everyone


First of all I would like to thank the engineers Tomelgy and Jlinton for their answer and I apologize for not responding sooner because I did not receive the notification an also this website Medwrench for giving me the opportunity to ask technical questions and sharing technical problems, But I think I did not explain the problem enough, So when I press the power button it beeps once then the screen stays black, The codes 6531 and 6910 flash on the control panel for 5 seconds  then the machine turns off by itself 


Thank you for your cooperation and please email me here zakariaidali4@gmail.com



Equipment: Seimens ARCADISOrbic 3D C-Arm


   RE: All Categories Report Post
Mon Oct 07 2024Reply from Tomelgy
avatar placemark

Hi

I just came across your questions (Oct 7, 2024).

6531: Is Dose adjustment out of range/cal. Check Iris movement etc.

6910: Is Dose range. Check mag ranges are changing etc.

You could try to reset error using the '0' button on gen console.

See if errors appear with radiation release.

Check TV iris is moving, particularly during startup.

Check dose rate adjustments. In old systems the Low dose mag1 may not be attainable as the II gain has dropped off with age. You either set the same value as Mag0 or select default values on page bottom as a test.

Tom


Report Post