Need Humidity Limites for the Century V-120, JACHO says we must know this so we don't oper: said same in the wrong inverioment,,,,,,,,???? should be somewhere in the enviromental operational peramitures...
Humidity
Hi Scottde,
All I could find was a MANUAL ON THEIR SITE that says it's 10-90%. The manual is for the "Century" but doesn't specifically call it a Century V120. Though maybe the V120 is just for the voltage configuration?
Good Luck!
Andrew Whyte, CBET, Maple Tree Biomedical LLC
Report Post
v120 and v116 refer to the dimensions of the chamber(v120 = 20 x 20 x 38?, v116 = 16 x 16 x 26? )
Report Post
and..
from the ops manual
2.2.3 Utility Requirements
Hinged Door — 44 x 511 /2 x 751 /4" (1118 x 1308 x 1911 mm) Sliding Door — 70 x 57 x 751 /4" (1778 x 1448 x 1911 mm) Hinged Door — 44 x 631 /2 x 751 /4" (1118 x 1600 x 1911 mm) Sliding Door — 70 x 69 x 751 /4" (1778 x 1753 x 1911 mm) Hinged Door — 44" x 751 /2" x 751 /4" (1118 x 1918 x 1911 mm) Sliding Door — 70 x 81 x 751 /4" (1778 x 2057 x 1911 mm) 3800 lbs (1720 kg) 4200 lbs (1901 kg) 4700 lbs (2127 kg)
Controls 120 V, 2 A, 1-phase Vacuum Pump 208/240 V, 6 A, 3-phase, or 480 V, 3 A, 3-phase
• Water:
Pressure: 20 to 50 psig (1.4 to 3.5 bar) Temperature: 70°F (21°C) maximum Consumption: 15 gpm (57 lpm), peak
• Steam:
Pressure: 50 to 80 psig (3.5 to 5.2 bar) Consumption:
36" Sterilizer: 190 lb/hr (86 kg/hr) 48" Sterilizer: 255 lb/hr (116 kg/hr) 60" Sterilizer: 335 lb/hr (152 kg/hr)
2-3
Installation Verification
Operator Manual
129373-461 2.2.4 Environmental Conditions
Temperature: 50° to 90°F (10° to 32°C) Humidity: 10% to 90% noncondensing Pollution Degree: 2 Installation Category (Overvoltage Category): II A-Weighted Sound Power Level: = 85 dBA (maximum)
Report Post
trying to follow your original post writing, what specifically is JACHO asking?
typically humidity questions are related to the room, from JACHO's website
https://www.jointcommission.org/standards/standard-faqs/ambulatory/environment-of-care-ec/000001275/