Humidity

Need Humidity Limites  for the Century V-120, JACHO says we must know this so we don't oper: said same in the wrong inverioment,,,,,,,,???? should be somewhere in the enviromental operational peramitures...

scottie.dewolfe@crothall.com


   RE: Sterilizers/Autoclaves Report Post
Sun Oct 20 2024Reply from MapleTreeBiomed
MapleTreeBiomed

Hi Scottde,

All I could find was a MANUAL ON THEIR SITE that says it's 10-90%. The manual is for the "Century" but doesn't specifically call it a Century V120. Though maybe the V120 is just for the voltage configuration? 

Good Luck!


Andrew Whyte, CBET, Maple Tree Biomedical LLC
Report Post
Wed Oct 23 2024Reply from QuirkletonAntsy
avatar placemark

Hi Scottde, 

Have you already contacted their service center?


Report Post
Wed Oct 23 2024Reply from Chris-CBET
Chris-CBET

v120 and v116 refer to the dimensions of the chamber(v120 = 20 x 20 x 38?, v116 = 16 x 16 x 26? )



Report Post
Wed Oct 23 2024Reply from Chris-CBET
Chris-CBET

and..

from the ops manual

2.2.3 Utility Requirements


Hinged Door — 44 x 511 /2 x 751 /4" (1118 x 1308 x 1911 mm) Sliding Door — 70 x 57 x 751 /4" (1778 x 1448 x 1911 mm) Hinged Door — 44 x 631 /2 x 751 /4" (1118 x 1600 x 1911 mm) Sliding Door — 70 x 69 x 751 /4" (1778 x 1753 x 1911 mm) Hinged Door — 44" x 751 /2" x 751 /4" (1118 x 1918 x 1911 mm) Sliding Door — 70 x 81 x 751 /4" (1778 x 2057 x 1911 mm) 3800 lbs (1720 kg) 4200 lbs (1901 kg) 4700 lbs (2127 kg)


Controls 120 V, 2 A, 1-phase Vacuum Pump 208/240 V, 6 A, 3-phase, or 480 V, 3 A, 3-phase


• Water:


Pressure: 20 to 50 psig (1.4 to 3.5 bar) Temperature: 70°F (21°C) maximum Consumption: 15 gpm (57 lpm), peak


• Steam:


Pressure: 50 to 80 psig (3.5 to 5.2 bar) Consumption:


36" Sterilizer: 190 lb/hr (86 kg/hr) 48" Sterilizer: 255 lb/hr (116 kg/hr) 60" Sterilizer: 335 lb/hr (152 kg/hr)


2-3


Installation Verification


Operator Manual


129373-461 2.2.4 Environmental Conditions


Temperature: 50° to 90°F (10° to 32°C) Humidity: 10% to 90% noncondensing Pollution Degree: 2 Installation Category (Overvoltage Category): II A-Weighted Sound Power Level: = 85 dBA (maximum)


Report Post
Wed Oct 23 2024Reply from Chris-CBET
Chris-CBET

trying to follow your original post writing, what specifically is JACHO asking?

typically humidity questions are related to the room, from JACHO's website

https://www.jointcommission.org/standards/standard-faqs/ambulatory/environment-of-care-ec/000001275/


Report Post