I am getting an error N183 and N 185 proximal occlusion. When I try to back prime I get error E433 malfunction
Fri Nov 08 2024
Plumb errors
Sun Nov 10 2024Reply from MapleTreeBiomed

Hi JaciD,
E433 means the proximal pressure sensor has failed.The service manual says to replace the Mechanism Assembly. You can find the service manual in the documents section on the left. The sensor failing would easily cause your other errors.
Good Luck!
Andrew Whyte, CBET, Maple Tree Biomedical LLC