Taking very long to heat up (30 min) to 'green zone'

We have an old Pelton and Crane that used to work well despite infrequent use/cleaning/servicing. Lately it has started to take 30 minutes to reach the 'green zone' for temp and PSI, and will often go over the maximum values of the green zone. Then the door latch is super hard to open, but a bit better after we cleaned the whole autoclave thoroughly. Any idea if this is something that we could repair ourselves? Once it is hot, it still works ok, but the drying process while on 'vent' also takes about 10 minutes rather than the advertised 3-5 min in the manual.

We might be looking for a replacement if we cannot fix ourselves, as we have gotten quotes of around $600 with no guarantee that it can be fixed from repair services.

Many thanks for any insight!


   RE: Sterilizers/Autoclaves Report Post
Mon Mar 10 2025Reply from Tyler K
Tyler K

Hi, the issue with the unit heating up slowly and it being an inconsistent temperature is most commonly due to a bad/old heating element. The PSI going over the safe zone could be due to a variety of reasons, with the most common being a faulty door seal, or faulty pressure relief valve in my experience. Unless you have someone who is incredibly handy or has experience working on autoclaves or even home appliances (somewhat similar to dishwashers), this may be a tall task. The most difficult step may be finding parts for a unit of this age. I would be more than happy to assist further or my company also sells used and new autoclaves too. 


Report Post
Thu Mar 13 2025Reply from PC_DVM
avatar placemark

@Tyler K: could you please write your company and contact information in a message as somehow I cannot click through to your profile information!


Report Post
Fri Mar 14 2025Reply from MapleTreeBiomed
MapleTreeBiomed

Hey PC_DVM,

Are you saying that the vent process takes 10 minutes or that drying everything takes 10 minutes? 

The OCR and OCM have a multi part bellows assembly that may be worn out. RPI has the rebuild kit. The bellows is supposed to let steam out until around 200F then close but it it's work out it may not be closing until much later or just not sealing at all. That lets all the water boil away and the temps can then spike. 

Good Luck!


Andrew Whyte, CBET, Maple Tree Biomedical LLC
Report Post
Sun Mar 16 2025Reply from JV
JV

The Pelton Crane is too old, to invest you money and time.


Report Post
Sat Apr 26 2025Reply from Jim F
Jim F

The Pelton OCM was built between 1966 &1980. Most rugged autoclave ever built!!  But, keeping this model running will be quite challenging. 

Temp fluctuations: The oldest model used a  mechanical thermostat, (evident by a 12" long slender metal tube laying in the center of the chamber). If that's what you have, this unit is pre 1972. Those thermostats can fluctuate greater than 5' as they age.  Solution: replace/upgrade to an electronic thermostat (repair shop). 

If it has an electronic thermostat, (evident by a cooling block (heat sink) sticking out the back side of the unit, there could be corrosion on the PC board. Or, the temp sensor is going bad. Again, only an experienced repair shop can help you.

Slow heating: If the heat eventually goes up to the sterilizing range and can stay there - at least a half hour, the valves are ok.

A bit of warning: if the wiring is covered with a soft cottony insulation - that is asbestos, and you should just retire the unit.


Report Post
Sun Apr 27 2025Reply from Chris-CBET
Chris-CBET

From what information you've given,  it sounds like your bellows is failing or vent valve isn't allowing all of the air to escape. 


Report Post
Sun May 04 2025Reply from michael1234
michael1234

Aso check the voltage of the sockets if is to low, sometime happened from power line.


michael1234

Report Post
Sun May 04 2025Reply from Chris-CBET
Chris-CBET

Any update?


Report Post
Mon May 12 2025Reply from CMB7
CMB7

Buy a new or 2nd hand unit, sediment and heat issue will cost you more money & time than is reasonable.


Bio-Medical Technological Services: Compressed Gas Systems Specialist Hyperbaric Chamber Technologist Services in: English, Français & Español