Errror 12-34-1-0

Please help me decipher the errors 12-34-1-0

appear when working on a large focus

Equipment: Siemens mammomat fusion


   RE: Mammography Report Post
Wed Apr 16 2025Reply from Jlinton
Jlinton

The error code 12-34-1-0 on the Siemens Mammomat Fusion typically refers to an issue related to the X-ray tube control or high-voltage generator, specifically when using the large focus setting.

Let’s break down the code:

 Siemens Error Code Structure (typical format):

So for:

  • 12 – Likely refers to X-ray generator / HV issues

  • 34 – Possibly points to focal spot control or tube filament current regulation

  • 1 – May refer to focus selection or exposure triggering issue

  • 0 – Indicates first occurrence or default detail code


 What This Likely Means:

The system is encountering a problem switching to or operating with the large focal spot. This could be due to:

  • A faulty filament or focus selector relay in the X-ray tube housing

  • An issue with filament current regulation (too low or too high)

  • A communication issue between the tube head and generator

  • A failing high voltage generator component that struggles with high load


 Troubleshooting Steps:

  1. Power Cycle System

    • Fully shut down and restart the Mammomat.

    • Clear the error and retry with small focus to isolate the issue.

  2. Inspect Filament Voltages (if service mode accessible)

    • Check if filament current/voltage for large focus is within normal limits.

  3. Try Low mAs Exposure on Large Focus

    • If it works at low power but fails at higher, it's likely a tube or HV generator degradation issue.

  4. Check for Recent Maintenance

    • Has the tube been replaced recently?

    • Was the generator calibrated properly post-install?

  5. Check Cables and Connectors

    • Inspect HV cables to the tube for signs of arcing or wear.

    • Reseat all connectors from the generator to the tube assembly.


James Linton, MiM, PmP, Cmbb, AAMIF
Biomedical Engineering Technology

Report Post
Thu Apr 17 2025Reply from dyakov
avatar placemark

Thank you 


Report Post
Mon Apr 21 2025Reply from Ohmeda
Ohmeda

Greetings to all.

I have already encountered this defect and 95% of the time it was always the defective tube. I am not saying that these defect codes are the same, but I ask that you focus on the tube.