Tue May 20 2025

Memoskop

mi equipo me presenta los siguientes error, 2006, 4021 y 4024, me podrían orientar por favor?

Equipment: Siremobil Compac L

Translation: My device is displaying the following errors: 2006, 4021, and 4024. Could you please help me? Equipment: Siremobil Compac L

   RE: Mobile C-Arm Report Post
Wed May 28 2025Reply from Matheus Silva
avatar placemark

Could be a bad memoskop, if you replace it, it may solve the problem 


Report Post
Wed May 28 2025Reply from Tomelgy
avatar placemark

Check sg cable & pins or do you get errors with just cart & no main system connected.

You can get into memoskop service & format hdd to see if it recovers

I have had the hdd fail

Tom


Report Post
Wed May 28 2025Reply from outinleft
avatar placemark

You cannot get Siremobil Host errors when only the monitor cart is powered up, since the errors are generated by the D-1 CPU board (located in the c-arm part, not the cart) and that part of the Siremobil is not powered up when the monitor cart is booted up separately.


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Wed May 28 2025Reply from outinleft
avatar placemark

2006, 4021, 4024 are all serial communication errors between the D1-CPU and the Memoskop, most probably either a bad Memoskop, or a bad SG cable socket/wire. Reach out to us if you need either of those parts. We usually have them both in stock.


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Wed May 28 2025Reply from Tomelgy
avatar placemark

The question regarding the Monitor cart without main system is to double check if the memoskop is working independently.

If the Memoskop is occupied with errors the min system, host D1 could report false positive.

This why the first suggestion was to check sg cable.

You could also check the 8 segment display on D1 is rotating before radiation request.

It would be interesting to know if these errors appear during boot or when radiation release is request


Report Post
Wed Jun 04 2025Reply from Matheus Silva
avatar placemark

Check if memoskop stay on minimum mode when you turn it on, if it does, its a bad memoskop


Report Post
Wed Jun 04 2025Reply from Tomelgy
avatar placemark

It could linked to the license file.

Is the dongle connected or working.

Tom


Report Post