Error 12064

Error 12064: Internal error F239

Equipment: Ziehm Imaging - Vision RFD Hybrid Edition


   RE: Mobile C-Arm Report Post
Mon Jun 09 2025Reply from Neomed Imaging - Service Dept
Neomed Imaging - Service Dept

If the GUI is not up yet, and this is a true boot error. Then it is likely a problem with the imaging system. Try and boot the monitor cart without the XP0 cable connected. If the error still comes up, you will need to repair or replace the imaging system.


NEOMED IMAGING - SERVICE - Ziehm C-Arm Parts, Repairs, Sales - Phone: 951-742-7020 - www.neomed-imaging.com
Report Post
Wed Jun 11 2025Reply from The Ziehm Guy
The Ziehm Guy

I agree with everything that Neomed had stated. However, if it does boot with the C-Arm connected, you can go into the service screen after entering the password: vision, then go to the System Tab, at the top of the screen, in the drop down menu, choose Versions, then see what CAN modules are listed. This may tell you what your issue is.... if it's not the Image Memory within the monitor cart.

Feel free to call if you have further tech questions.

Surgi-Call - Ziehm and OrthoScan Service - Phone: 302.766.5225 - 

www.surgi-call.com


Report Post
Thu Jun 12 2025Reply from BillyTemple
avatar placemark

Thank everyone for trying to help.  I did contact Ziehm.  The issue was they tried to change the protocol while imaging.  So, operator error.


Report Post
Fri Jun 13 2025Reply from The Ziehm Guy
The Ziehm Guy

Glad everything is working ok!