Wed Aug 20 2025
error 450 on th mammomat 3000 nova? The unit uses agfa cr cassettes. Thank you
Good day. Any ideas to clear error 450 on mammomat 3000 nova? The error rfers to the AEC. I have manually increase the kV up to 30kV with 4cm compression still no luck
   RE: Mammography Report Post
Sun Sep 28 2025Reply from deesmithk
Err 450 is for dose monitoring, basically either the dose is very low or non-existent. You could have a fuse blown, a bad x-ray tube or a bad board, a bad connection in the generator or at the tube. To get more info on this unit go to Franks Hospital equipment/x-ray equipment/medical imaging.