Equipment: UNICO - PowerSpin DX
ERROR E-4
RE: Centrifuge Report Post

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.
Just be sure to tell them you were referred by MedWrench
Report Post

E-4 on a PowerSpin DX usually means the controller isn’t getting a valid lid/door or rotor home signal at startup — the machine refuses to run for safety. Could be a simple switch/wiring issue or a failing position sensor/drive board.
Quick checks (do these first):
Power-cycle completely (main off ? wait 30s ? on).
Open/close lid several times and watch/feel the lid-lock engage.
Inspect lid interlock microswitch and wiring for loose plugs, corrosion, broken pins.
Manually check rotor can move freely (power off) — no jams or debris.
Check main fuses and the motor/drive supply voltages if you can safely access them.
If there’s a service/homing routine in the menu, run it and note where it stops.
James Linton, MiM, PmP, Cmbb, AAMIF
Biomedical Engineering Technology
Report Post
Thank you for your response. I will do the checks you recommended first. Also, I will check the start circuit of the motor, because the centrifuge doesn't show the error if you give a little assistance to the rotor to start.