Thu Nov 13 2025
Switch for moving table up/down
Switch for moving table up and down is not working, in case press switch for moving down it is not responding at all, when press switch up it not moving at all, but when release switch it shake table a bit, like it would like to move up or the mechanism is stuck. have any the same experience and how this problem can be solved? thank you for any help.
Equipment: ZEISS VISULAS 532s
Tue Nov 18 2025Reply from Jlinton
Jlinton
Most common fix
- Replace the lift motor start capacitor or unstick a stuck limit switch.
Solves most of the cases I’ve seen on these tables.
Wed Nov 19 2025Reply from G.E.
Mr. James thank you for response, do you have do you have technical manual for this table? I does not find it on the net, If you can share it I would be greatful.
Wed Nov 19 2025Reply from Jlinton
Jlinton
I do not sorry
Wed Nov 19 2025Reply from Laserdude
If you have a dual leg on the slit lamp table or do you have a single leg on the table top to the floor?
Thu Nov 20 2025Reply from G.E.
Table has single leh from the table to the floor, it is from document for laser but in the document it is not technical manual of this table.
Thu Nov 20 2025Reply from Chris-CBET
Chris-CBET
is it making any noise? maye your switch or switch connection is bad
Fri Nov 28 2025Reply from frobles8
There is not manual of the table mechanism.
You must disesambly the table and look inside of the elevator mechanism.
I had two times a problem like you and after disesambly the elevator mechaniosm, i realize that the limit switch inside were damage or broken.
You have to replace it.
Good look G.E.