Planmed Sophie Classic shows error code 01 when pressed exposure switch

Planmed sophie classic gives error code 01 when pressed expose switch.  Any suggestion on how to rectify this.  thanks.


   RE: Mammography Report Post
Mon Dec 01 2025Reply from Jlinton
Jlinton

Err-01 on a Sophie Classic is basically “no HV/MG start” — the tube isn’t getting the all-clear to fire.

In plain terms: one of the interlocks isn’t closing.

Quick checklist (in the order these usually fail):

Door interlock switch – most common. Dirty, loose, or out of alignment.

Arm/travel limit switches – the unit won’t expose if a limit switch isn’t made.

Compression paddle microswitch – must be fully latched.

Prep (rotor) signal – bad coil or loose wiring on the prep relay.

HV generator enable cable – reseat both ends; they loosen with age.

Nine out of ten times Err-01 is a door switch or compression switch acting flaky. Fix the bad interlock and the code clears immediately.


James Linton, MiM, PmP, Cmbb, AAMIF
Biomedical Engineering Technology