Tue Nov 25 2025

E10 powers off

When I turn on my E10 i reaches the GE startup screen with the progressbar. Then it stops - a messagebox apperas with an error and it turns off. I never hear the relays clicking.

If I turn the machine off at the outlet and wait until it compleatly turns off  and then turn it on it starts normally. I do get an battery deteroiated error when booted. If I try to restart (without turning it off at the outlet) it won´t start op again? 

Is this a common behaviour when the battery is dead?

Later:

I got a battery communication error - when I pressed the battery icon at the bottom. When I turned the machine off and tries to start it again it stops at the progressbar whatever I do and shows the error message and turns off.


Ok, the culprit was the batteries - the machine for some reason won´t start if there is a communictaion error with the batteries. Removed them and now it starts - every time.

Have ordered new ones. 



   RE: Ultrasound Systems Report Post
Tue Nov 25 2025Reply from ><}}}}> BKC
avatar placemark

I recommend unplugging the machine and disconnecting all batteries, then letting it sit with no power applied for at least 30 seconds. After that, try powering it back on. If the issue persists, please pull the EMB and check the BIOS battery voltage.


I have seen a dead bios battery cause a number of issues with the E10 


Report Post
Tue Nov 25 2025Reply from ><}}}}> BKC
avatar placemark

the correct bios battery is BR2032, not CR2032,  you can use a CR2032 for testing thou 