Defrost cycling causing condensation inside chamber and then causes ice buildup

Hi,

We have recently noticed ice building up on the knife holder and carrier to the point that sometimes you cannot adjust the angle of the carrier or the forward and lateral movement.  We have also seen frozen ice droplets in the chamber.  We defrosted the cryostat but the problem persists.  We then watched the defrost cycle and noticed that the chamber gets very warm and a lot of condensation occurs on the window and inside the chamber.  The display shows H - 42, which I believe indicates the temperature of the evaporator. Once the defrost cycle ends there is still a lot of condensation in the chamber and this appears to be reason we see freezing once the cryostat cools down.  Any ideas on why this happens and how to fix it are appreciated.  This started about a month ago and has never happened during the past 12 years that we have used the cryostat.  Thanks..

Equipment: Microm HM-525m Cryostat


   RE: Cryostat Report Post
Mon Dec 01 2025Reply from TECHTE
avatar placemark

During the defrost cycle, does the H-42 ever change or does it say that the whole time?  Also, what is the display saying temp inside the cryostat?


tech.te.usa@gmail.comPowered by Froala Editor
Report Post
Tue Dec 02 2025Reply from MapleTreeBiomed
MapleTreeBiomed

How is the seal around the window? It may be sucking in humid room air when cooling down and condensing inside the chamber. Much like a fridge with a bad door seal will grow a puddle or the freezer fills up with ice.


Andrew Whyte, CBET, Maple Tree Biomedical LLC
Report Post
Tue Dec 02 2025Reply from FV
avatar placemark

Thank you for responding.  I wanted to go back and observe a full defrost cycle but since it was lasting too long, greater than 45 minutes,  I interrupted it.

As soon as the defrost cycle starts, within the first minute H-42 shows up on the temperature display and remains there throughout the defrost cycle.

After the first 5 minutes the temperature in the chamber as measured by my own thermometer was -11C.

After 21 minutes it was +2C.

After 25 minutes it was +4C.

At this point condensation started appearing on the window and on the microtome surface inside.

The window insulation looks secure since we do not see any issues once the cryostat cools down to -20C, there is no frost buildup inside or around the edges.  





Report Post
Tue Dec 02 2025Reply from TECHTE
avatar placemark

Ok, it would appear that the temp sensor for the defrost cycle is broken, thus the defrost cycle is running too long and it is causing this ice and frost issues.  This sensor is difficult to obtain, if you need assistance on that please let me know.


tech.te.usa@gmail.comPowered by Froala Editor
Report Post
Tue Dec 02 2025Reply from FV
avatar placemark

Can you tell me where the sensor is located and how to get access to it?  This issue started shortly after someone did preventative maintenance on the cryostat so maybe the sensor has fallen off or the wire has been damaged.


Report Post
Wed Dec 03 2025Reply from TECHTE
avatar placemark

It is highly unlikely that this happened during the PM is the sensor is not easily accessible. It is located in the evaporator and requires taking off the top.  I recommend that a service tech do this repair.


tech.te.usa@gmail.comPowered by Froala Editor
Report Post
Wed Dec 03 2025Reply from TECHTE
avatar placemark

Also, this sensor is not readily available as parts have been discontinued for this model.  Also, it is not a generic sensor like some models.  If you need assistance with the repair or obtaining the part, please email me.


tech.te.usa@gmail.comPowered by Froala Editor
Report Post
Wed Dec 03 2025Reply from FV
avatar placemark

Thank you for all your help.  I have a repair tech coming in next week and will discuss it with him and let you know if we need assistance with obtaining the part.