Hgb lamp failure

My sinothinker SK9000 keeps showing this error HGB Lamp failure and the result is inconsistent 

What should I do please 

Equipment: Sinothinker SK9000


   RE: Hematology Analyzer Report Post
Mon Dec 01 2025Reply from Jlinton
Jlinton

Replace the HGB lamp module – this fixes it 90% of the time. They’re consumables and don’t last long.

Check the lamp connector – make sure it’s fully seated; loose connection = flicker = inconsistent results.

Clean the optical path – dust or cuvette residue can also mimic a weak lamp.

Run a full optical calibration after replacing the lamp.

If the error stays after a new lamp and calibration, the HGB photodiode board is your next suspect, but that’s far less common.


James Linton, MiM, PmP, Cmbb, AAMIF
Biomedical Engineering Technology