ERBE - Erbotom ICC 200 don`t have full energy
RE: Electrosurgical Unit (ESU) Report Post

how are you testing? are you using a simulator or actually using it on a patient?
when you say you can't regulate, do you mean you can't adjust the output or you're not getting any output
Report Post
I used a piece of lard as test. The device work but power 10w or 100w the same, i don't see different. Bipolar or monopolar - the same. Maybe something limits power?
Report Post

There's no way to tell what the output is with that. Where's your return electrode/REM plate?
Report Post

There's no way to tell what the output is with that. Where's your return electrode/REM plate?
Report Post
NESSY is OK, i guess. I use non-split electrode. May be this electrod is not from this device i don`t know
Report Post

If the rem pad is too far from the cutting site, the resistance is increased and this will affect your output.
Report Post