ERBE - Erbotom ICC 200 don`t have full energy

Hi! I have an Erbe ICC200 device. No errors. The power of the device is small and don`t regulated. Can somebody help me? I am an engineer, not a surgeon.

   RE: Electrosurgical Unit (ESU) Report Post
Fri Dec 19 2025Reply from Chris-CBET
Chris-CBET

Are you using the right load specs?


Report Post
Fri Dec 19 2025Reply from serdgik
avatar placemark

I guess yes. It works but i can't regulate power


Report Post
Fri Dec 19 2025Reply from Chris-CBET
Chris-CBET

how are you testing? are you using a simulator or actually using it on a patient?

when you say you can't regulate, do you mean you can't adjust the output or you're not getting any output


Report Post
Sat Dec 20 2025Reply from serdgik
avatar placemark

I used a piece of lard as test. The device work but power 10w or 100w the same, i don't see different. Bipolar or monopolar - the same. Maybe something limits power? 


Report Post
Mon Dec 22 2025Reply from Chris-CBET
Chris-CBET

There's no way to tell what the output is with that. Where's your return electrode/REM plate?


Report Post
Mon Dec 22 2025Reply from Chris-CBET
Chris-CBET

There's no way to tell what the output is with that. Where's your return electrode/REM plate?


Report Post
Mon Dec 22 2025Reply from serdgik
avatar placemark

NESSY is OK, i guess. I use non-split electrode. May be this electrod is not from this device i don`t know


Report Post
Mon Dec 22 2025Reply from Chris-CBET
Chris-CBET

If the rem pad is too far from the cutting site, the resistance is increased and this will affect your output. 


Report Post
Tue Dec 23 2025Reply from serdgik
avatar placemark

no, the rem pad is near to the cutting site. I know about increasing resistance


Report Post