Sat Dec 20 2025

Mamografo

Tengo estos errores 25:8 y 25:11 error de captura de imagen y 25::11 filament calibration is recommended

Equipment: Hologic selenia dimensions

Translation: I'm getting these errors: 25:8 and 25:11 image capture error, and 25:11 filament calibration is recommended.

   RE: Mammography Report Post
Fri Jan 02 2026Reply from Jlinton
Jlinton

On Hologic Selenia Dimensions, 25:8 and 25:11 are X-ray chain–related image capture faults.

  • Run Filament Calibration from service mode first
  • If calibration fails or the errors return, suspect tube aging, filament instability, or the HV tank
  • Check recent line voltage and confirm no intermittent power issues
  • If exposures are inconsistent or noisy, this usually ends in tube replacement, not detector work

James Linton, MiM, PmP, Cmbb, AAMIF
Biomedical Engineering Technology

Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.