Fri Jan 09 2026

Self test failure

The device works perfectly. I have a fault at start up: HCU Supply. 00110006.

What can be the cause of this fault and how can I repair.

Equipment: MAQUET - HCU 40


   RE: Warming/Cooling Unit Report Post
Fri Jan 16 2026Reply from Jlinton
Jlinton

HCU Supply 00110006 on an HCU 40 is a power-supply monitoring fault.

Most common causes:

  • Weak or failing internal PSU
  • Loose PSU or backplane connector
  • Failing voltage sense line (supply is present but out of tolerance at startup)

What to do:

  • Power off, reseat PSU and internal connectors
  • Check PSU output voltages under load
  • If it repeats, replace the power supply module

James Linton, MiM, PmP, Cmbb, AAMIF