Sun Feb 15 2026
irrus 600 Ceiling Patient Lift Trouble shooting
Hello
Need some help, bought a used lift online and get it all mounted and hooked up. All features work but the down feature. It does work pulling the emergency strap but after emergency strap is used can’t up feature without turning it on and off. Seems like a “safety” feature that I can’t figure out. Anyone have and experience trouble shooting this brand
Equipment: cirrus 600 Ceiling Patient Lift
   RE: Cath/Angio Report Post
Tue Feb 17 2026Reply from MapleTreeBiomed
MapleTreeBiomed
Hey Parentfiguringitout,
Does the down button on the unit (not the hand control) work? I'd test it with and without the hand control attached. Either way I would suspect the hand control first.
Tue Feb 17 2026Reply from Parentfiguringitout
Hello thank you for your reply. Unfortunately neither button work on the remote or the device itself for the down feature. What would be the next trouble shooting step. All features work but this, including the emergency down feature. I greatly appreciate your time
Tue Feb 17 2026Reply from Jlinton
Jlinton
Since:
  • Up works
  • Emergency down works
  • Normal down does NOT work (both panel and hand control)
This means the motor and actuator are fine. The problem is in the down control circuit, not the lift motor.
Most likely causes:
  • Faulty down limit switch
  • Stuck safety interlock after overload
  • Control board relay for “down” failed
  • Anti-crush / slack strap sensor triggered
Next steps:
  • Check for a bottom limit switch stuck open
  • Inspect down button wiring continuity
  • Look for a small micro-switch near the strap drum
  • Reset fully (power off, remove battery 5–10 min)
Wed Feb 18 2026Reply from Chris-CBET
Chris-CBET
When you press the button,  ate you heating any clicking,  grinding,  etc?
Might just be a touchpad issue
Sun Feb 22 2026Reply from Parentfiguringitout
Hello
This red light blinks instantly after letting go of the the down bottom, no noise
All other times these are the three green lights on, cannot load picture. But the three green lights on the right are on. Cannot load picture