When the machine is turned on after the computer boots, the console monitor displays the system messages with the codes ASP_GEP_E_5144, ASP_MSP_4152 and ASP_MSP_5203. Can anyone help me?
Generator D21 error ASP_GEP_E_5144 troubleshooting
Thank you for your reply, Michael. Yes, I see the documents. Do you have troubleshooting guide for these codes?
Report Post

You are right Felix, the documents and Forums are not enogh to resolve the problem but I found the complete Operate Manual from your machine and maybe you will find there something about these codes.
How do I send to you? or you have already this with 478 pages?
michael1234
Report Post

After check this manual and do not find the problem you can connect via online with Siemens and they will give immediate assistance and this assistance it is in the manual how you will do it Felix.
michael1234
Report Post
Error GEP-5144 is internal hardware error.
Check basic stuff like power supply voltages.
Shielding
D21 might be faulty.
Check LED status
It could be a Tube Filament failure.
Could be a focus change error.
Check cables to x-ray tube.
Tom
Report Post
Thanks for the information, Tom. Do you have any information on what the status of each LED means?
Report Post
D21 leds.
V80. (Red) D21 not initialised
V73. (Red) X-ray tube voltage, kv-max + anode
V72. (Red) X-ray tube voltage, kv-max - cathode.
V70. (Red) Over current filament inverter.
V69. (Red) I-max-kv. Over current in inverter.
V68. (Red) I-max-rac. Over current in rac.
V59. (Green) Swr-D21. Generator ready for radiation.
V60 (yellow) swr, inverter on.
V57. (Green) xray-on.
V54. (Green) swr-fil. Filament active.
V50. ( red) fault in power supply.
V49. (Red) tube Over pressure switch active.
All these and more are on the circuit diagrams.
You coukd download the circuit from siemens web site.
Tom