Tue Apr 07 2026

Connect AC Plate Error

Good day,

I am having issues with the power for the table hence the connect AC plate error. The fuses keep blowing, I am out of 4A fuses, so I tried yth3.15A and they blow instantly. I no longer have the service manual due to unexpected circumstances, so I am asking for any help and tips. I am located in Iraq, so we are using 230-250V. The unit has recently moved under not ideal situation, so I am not sure if something got damaged during the relocation. I am trying to acquire T4AH fuses that are on the data plate, but I feel like there might be something else wrong with the table.

Thank you in advance for any assistance provided.


   RE: Surgical Table Report Post
Tue Apr 07 2026Reply from MapleTreeBiomed
MapleTreeBiomed

Hey KBRMET,

Do you have a multimeter? I don't have the service manual but something is obviously shorted. I would remove the bottom housing and start looking for anything obvious. A loose screw resting on the power board for example. It's not likely to be a motor so you are down to internal wiring and the boards. 


Andrew Whyte, CBET, Maple Tree Biomedical LLC
Report Post
Tue Apr 07 2026Reply from Chris-CBET
Chris-CBET

You've got a dead short somewhere. Better isolate that b4 you waste any more fuses. 


Report Post
Tue Apr 07 2026Reply from KBRBMET
avatar placemark

Thank you Andrew and Chris, I will try my best to find it.