The equipment won't turn off using the power switch

I got a unit for repair that won't turn off. It starts up fine, but when you hit the power switch, it just keeps rebooting in a loop until the battery dies. I tried swapping in a power supply from a working machine to test it. Not only did it not fix the problem, but it seems to have fried the good power supply too—now that machine is broken as well.

Equipment: Dräger Savina 300


   RE: Ventilator Report Post
Tue Apr 07 2026Reply from Chris-CBET
Chris-CBET

did you check the power switch?


Report Post
Tue Apr 07 2026Reply from ArturGaspar
avatar placemark

I didn't replace the part, but I swapped the wires to test it; the two top pins were moved to the bottom and the bottom ones to the top.
And the problem remained the same.


Report Post
Tue Apr 07 2026Reply from Chris-CBET
Chris-CBET

if you disconnected the wires from the power switch and it didn't turn off...that sounds like a short somewhere and if the power supply got fried, that would support a short. 


Report Post
Thu Apr 09 2026Reply from ArturGaspar
avatar placemark

I didn't leave the leads unhooked; I only reversed the polarity at the switch. I swapped the switch for a new one today, but the issue remains. I suspect the power supply unit (PSU) is faulty, but I can't locate the circuit diagram to perform proper troubleshooting.


Report Post
Tue Apr 14 2026Reply from Paul D
avatar placemark

the power switch is actually a software switch, not a true power switch. The Savina shuts down via an on screen menu. Power supply is a self contained module.


Report Post