Thu May 07 2026

Temp tolerance

Trying to find temp tolerance for Gen 6 blanket warmer

Equipment: STERIS - AMSCO QDJ06


   RE: Warming Cabinet Report Post
Fri May 08 2026Reply from MapleTreeBiomed
MapleTreeBiomed

Hey St3v3,

I found this in the generic manual Steris has on their site for the Amsco cabinets. 

"Air temperature near fan inlet should be within 5°F (3°C) of display." I suppose +/- 5 degrees F. Though somewhere I swear I saw 3% but I just cant find that again. 

Good Luck! 


Andrew Whyte, CBET, Maple Tree Biomedical LLC
Report Post
Sun May 17 2026Reply from Chris-CBET
Chris-CBET

5th Gen was +- 5 degrees f, so it's most likely what Andrew stated. I don't believe the controls have really been updated b/w 5th and 6th generations so it's not likely the calibration has changed


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree