NIBP air leakage

Greetings Engineers The above mentioned patient monitor display "air leakage"error everytime NIBP is carried out. I changed the NIBP accessories with the one that is working but we still have the same results"air leakage". Please what can I do? I am a biomedical engineer from Cameroon working with Mbingo Baptist hospital. Thank you Waiting

Equipment: Datex-Ohmeda cardiocap/5


   RE: Telemetry-Central Station Report Post
Mon Jun 29 2026Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

Ampronix

Prescotts

Sage Services Group

 

Just be sure to tell them you were referred by MedWrench 


Report Post
Fri Jul 10 2026Reply from jdeater
avatar placemark

You will probably need to crack open the case on the unit. find the NIBP tubing and carefully trim each end where they connect to nipples. The tubing has most likely relaxed over the years and is not tight on the nipples anymore. I simple trim of a half cm or a little more and refit them on should remedy your leak issue.


Report Post
Mon Jul 13 2026Reply from BrentS
BrentS

Typical leaks are usually the cuff, hose, hose o-rings or a combination of. Use Snoop or soapy water to look for leaks or swap out with a known good working set.

If none of the above are the issue, you'll have to check for internal leaks. Usually the pump, it's tubing or both have will have leaks and require replacement. This monitor is very old and past EOL. I would suspect the pump has leaks. You may have a hard time finding a replacement pump. If you do replace the pump, you're required to perform a pump calibration procedure.


Brent SilviusAnesthesia TechnicianBiomedical Sales and ServicesMercuryMed.com