Equipment: Cutera - XEO Platform
XEO Cooling System Failure -- Pump and Filter Replacement / Service Technician Needed
RE: Aesthetic Laser Report Post
E322 could be caused by a few problems: if your IPL/dummy plug is not properly connected, a faulty pressure switch or a water level sensor.
1. Make sure that the connector is fully seated.
2. Pressure switch can be adjusted with a small set screw if accessible.
3. Replace a water level sensor that's inside a reservoir.
It is plausible that a water pump is malfunctioning, but in my experience it is rare. New DI and water filters are recommended every 12 months of service.