system vented or drive gas missing

Translation of original post:

When I try to test compliance status I get this: system drive ventd gas or missing ... How do I resolve?

 

Oringinal post:

Cuando intento hacer el test de compliance me aparece ese status: system ventd or drive gas missing... Como lo resuelvo


   RE: Anesthesia Gas Machine Report Post
Sat Aug 17 2013Reply from Apola
avatar placemark

Creo que es un problema de presion en el sistema, verifica si la prueba de fuga.

1 Thread down the O2 cell and press it down into its mount 2 Replace the breathing circuit 3 Cap the sample line leading to the agent monitor 4 Cap the Recirculation Port (if installed) located behind the auxiliary O2 flowmeter 5 Leaking O-rings at ypiece seal 6 Inspect bellows dome Oring 7 Check Neg. Pressure relief valve membrane is present 8 Tighten INSP and EXSP ring   o tambien

reason                                                                    solution

O2 cell O-ring                                                       Replace leaky O2 cell diffuser                                                                                 O-ring Breathing circuit leaks                                        Replace breathing circuit Loose valve rings                                                Hand tighten Leaking bellows or dome                                  Replace bellows Leaky docking station O-ring                             See Section 4.0 for docking (C, F or G)                                                              station O-ring locations   2.9 psi regulator out of cal (too                         Check/calibrate Pressure low)

Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree