HGB Reference

High HGB when running controls. HGB Reference does not adjust above 1795. HGB flow cell cleaned. What recommendations would you suggest?


   RE: Clinical Chemistry Analyzer Report Post
Tue Jul 24 2012Reply from xtremeboarder
avatar placemark

Jut enter the calibration screen then enter factor and adjust the factor to something lower. Example: 1.03 change to 1.00. Or u can re calibrate the system. I dont know what you mean by the reference wont adjust above 1795 ? If you dont have a calibrator use the normal control means to calibrate.

 

7 Years Cell Dyn Field Service


Report Post