High HGB when running controls. HGB Reference does not adjust above 1795. HGB flow cell cleaned. What recommendations would you suggest?
Mon Jul 16 2012
HGB Reference
Tue Jul 24 2012Reply from xtremeboarder

Jut enter the calibration screen then enter factor and adjust the factor to something lower. Example: 1.03 change to 1.00. Or u can re calibrate the system. I dont know what you mean by the reference wont adjust above 1795 ? If you dont have a calibrator use the normal control means to calibrate.
7 Years Cell Dyn Field Service