Fri May 13 2011
interference on ECG monitor
is there a way to eliminate interference
   RE: Patient Monitoring Report Post
Fri May 13 2011Reply from GaryBiomed
GaryBiomed
Not sure what monitor your having problems with, but mindray/datascope makes a red cable for ESU interference. I'm sure other manufactures make one also
Fri May 13 2011Reply from JayFromBiomed
Are you saying that you have a noisy ECG trace or that your entire display is noisy? Noisy trace is most commonly dry electrodes or poor patient prep, followed by bad cables. Usually not the monitor itself.