Urgent Product Correction Notice
Tue Jun 14 2011
Beckman Coulter AcT 5 AL Correction Action
Tue Jun 14 2011Reply from MedWrenchManager

What is the correction notice? If you have a document to share, you can post to the equipments product page or email me at kaylee@medwrench.com to post on your behalf.
Report Post
Tue Jun 14 2011Reply from viper1947

The correction notice is an Urgent Product Correction from Beckman Coulter on this system dated 2/23/2011. I would like a copy of this as I gather Beckman Coulter sent this to all AcT 5 AL customers warning them to no longer use the autoloader or cap piercer and to open the samples one at a time and run through the probe due to false high RBC, HCT, PLT & MPV results.