Wed Jun 15 2011
OR table with intermittent issues
Does not recognize AC. Will not switch to battery when ac is removed. No charge indicator. All symptoms are intermittent. Thinking the relay control box. Box contains two PCBs, anyone know where the boards can be purchased individually?
Wed Jun 15 2011Reply from Yanez
SKYTRON? (616)656-2900 or Northern scientific ( Refurbished or after market) not sure if they carry skytron parts.
Wed Jun 15 2011Reply from ruforjc2
ruforjc2
Thanks Yanez!
Wed Jul 27 2011Reply from Michael Sturgis
Did you get this taken care of, this is a common problem on the 3600b table. What has to be done is , rotate the incoming power voltage selecter pot. They get caroded and the table wont recognize the ac. Just turn it back and forth a couple times and its fixed.
Michael Sturgis, Sr.Biomed Tech.
Biomedical Engineering Department
Mercy Hospital Cadillac
sturgim@trinity-health.org
Fri Jul 29 2011Reply from ruforjc2
ruforjc2
Thanks Mike, Our issue ended up being the relay control box....ouch! Really appreciate the insite, going forward it may well save us some dollars! Thanks again, Mike