Thu Aug 02 2012

LAMP CHECK ERROR

Dear sir,

In my instrument, otheroptions, lamp check is giving a value of high ie 171.966 and no absorbace values are getting but only 0.000 always.  How to correct it? normally lamp check value comes around 43.000.  Is there any problem with photometer board? kindly revert back

 

tojo

 


   RE: Clinical Chemistry Analyzer Report Post
Thu Sep 06 2012Reply from sachidan
avatar placemark

check powersupply to photo diode board

171 is literally the infinite value that means , photo diode board problem

 


Report Post
Mon Sep 10 2012Reply from tojojerome
avatar placemark

Dear sir,

 

do you have service manual for BTS series of instrumets? If you can provide this will be very much helpful for me


Report Post
Sat Nov 02 2013Reply from tojojerome
avatar placemark

Dear Sir,

 

Problem solved....it was due to ribbon cable connectivity problem with photodiode.

 

thanks


Report Post
Sat Nov 23 2013Reply from Van Tien
avatar placemark

well done . i had a same your problem  . thanks for your comments.


Report Post