Wed Aug 08 2012

failure

1.lower screen doesn't come on/power up

2.the mourse/cursor knob doesnt work


   RE: Ultrasound Systems Report Post
Thu Aug 09 2012Reply from GaryBiomed
GaryBiomed

You can get the service manual here

http://www3.gehealthcare.com/en/Support/Support_Documentation_Library

 

You should contact Trisonics

1-717-939-6860

or

Unisysn

1-(303) 384-3246

for tech support


Report Post
Mon Aug 13 2012Reply from summit_mike
avatar placemark

I can definately help you with this problem. Give Mike @ Summit Imaging a call - 1-866-586-3744.

I have a few tricks we can try out.

mike@mysummitimaging.com


Report Post
Mon Aug 13 2012Reply from summit_mike
avatar placemark

I can definately help you with this problem. Give Mike @ Summit Imaging a call - 1-866-586-3744.

I have a few tricks we can try out.

mike@mysummitimaging.com


Report Post
Tue Aug 14 2012Reply from ekamanda
avatar placemark

Thanks mike will call

 