Toshiba PVT 575MV

I keep getting an error when using the Toshiba PVT 575MV:

"Mechanical Transducer Error: Please select another transducer or reboot the system and then select mechanical transducer."

What do you think is the issue and is it repairable?


   RE: Ultrasound Systems Report Post
Mon Aug 13 2012Reply from CCMS, Inc.
CCMS, Inc.

Hello Hung Tran,

Either the probe is bad or the board that controls it is bad.

To find out which one, I would try the probe in a different probe port on the system. If it still fails, try another system. If the problem follows the probe, the probe is definitely bad.

Regards,

Christopher Turner



C&C Medical Solutions, Inc.