Screen will not calibrate... monitor will not accept touch of 2nd target in Cal Mode. I replaced the bezel and touch sceen with Philips parts, same result.
MP30 Touch Screen Calibration

Probably needs a new touch panel. I would call Philips support at 1-800-722-9377 and see if they have a better idea.
Answer provided by Jeffrey E., Biomedical Equipment Technician at Sisters of Providence Health System.
Report Post

Philips have advised the HIF Board should have also been replaced. I ordered one and will pass on the results when it is installed.
Report Post

HI.
Have you ever log in to service mode to fix this problem. I think this function is disable. Please try below steps:
To permanently disable touchscreen operation:
a. Enter Service Mode.
b. Select Main Setup
c. Select User Interface
d. Change the Touch Enable selection to no.
To re-enable touchscreen functionality change the Touch Enable selection to yes.
Thanks
Report Post

Hi, Tried your suggestion - unfortunately, did not work. But thank you for your help.
Darryl
Report Post

i had the same problem.
I removed power cord and batteries.
Pressed the power button for 10 seconds.
I then put the batteries in connected power and calibrated the screen.
Works great
Report Post