Tue Aug 21 2012

On GE Procare - Error codes

Hey All !,

I have a GE procare that gives Error 950.  manul does'nt say anything except to contact customer service(for errors in the range of 900 and above). Has anyone had such an experience and was it serviceable in-house in the Biomed shop?

Murali

 

 


   RE: Vital Signs/Multi-Parameter Report Post
Wed Aug 22 2012Reply from MedWrenchManager
MedWrenchManager

Which model number do you have?


Report Post
Wed Aug 22 2012Reply from MURALI
avatar placemark

Front Panel says PROCARE. Serial Number plate says: DINAMAP DPC 200X - EN.


Report Post
Wed Sep 18 2013Reply from NakaDeshi
avatar placemark

Quote MURALI

Hey All !,

I have a GE procare that gives Error 950.  manul does'nt say anything except to contact customer service(for errors in the range of 900 and above). Has anyone had such an experience and was it serviceable in-house in the Biomed shop?

Murali

 

 

THIS ERROR CAUSED BY PUMP MOTOR OVER CURRENT CHECK THE PUMP MSNIFOLD (GREEN BLOCK). CONNECTED TO MOTOR FRONT. INSIDE THERE'S CHECK VALVES AND NET MESH. MESH MIGHT BE DIRTY. CLEAN THIS ON BOTH INPUT AND OUTPUT AIR. TRY THIS AND SEE IF IT WORKS.