Accuracy Test

I ran the accuracy test as follows rate: 200mL/hr and VTBI: 20mL and the pump delivers correctly every time. When running the accuracy test according to the service manuals set rates there is always an occlusion within the first 3mL. Has anyone run into this issue or have a idea for a solution?


   RE: Infusion Pump Report Post