Please can some body help me this error during bootup
Mon Aug 22 2011
E 10
Thu Aug 25 2011Reply from GaryBiomed
[url=http://www.reichert.com/contact.cfm]http://www.reichert.com/contact.cfm[/url]
From the user manual - [url=http://doclibrary.com/MSC167/PRM/BILIRUBINOMETER2432.pdf]Link Here[/url]
E10 - Problem cuvette not in sample path
Solution1 - Turn instrument off and on, curvette will rotate back to initial position. Seat cuvette completely in holder
Solution2 - Motor problem; contact service