E 10

Please can some body help me this error during bootup

   RE: Bilirubinometer Report Post
Thu Aug 25 2011Reply from GaryBiomed
GaryBiomed
[url=http://www.reichert.com/contact.cfm]http://www.reichert.com/contact.cfm[/url] From the user manual - [url=http://doclibrary.com/MSC167/PRM/BILIRUBINOMETER2432.pdf]Link Here[/url] E10 - Problem cuvette not in sample path Solution1 - Turn instrument off and on, curvette will rotate back to initial position. Seat cuvette completely in holder Solution2 - Motor problem; contact service Reichert Analytical Instruments Reichert, Inc. 3362 Walden Ave. Depew, NY 14043 USA Toll Free: 888-849-8955 Phone: 716-686-4500 Fax: 716-686-4555 email: ai@reichert.com www.reichert.com Reichert GmbH Hubertusstrasse 2 D-82229 Seefeld Germany Tel: +49-8152-993530