Blanket Warmer Temp

Where can I find guidance for the suggested temp of a blanket warmer?


   RE: Warming Cabinet Report Post
Fri Sep 16 2011Reply from MedWrenchManager
MedWrenchManager

Hi Robert,

According to Blickman, the suggested range would be 90-160 degrees fahrenheit. For the suggested temperature, you would have to contact the joint commission.

Hope that helps!


Report Post
Tue Nov 01 2011Reply from Jamie H
avatar placemark

We use the recommendations of ECRI for 110 degrees Fahrenheit for blankets and 104 degrees Fahrenheit for fluids.


Report Post
Wed Nov 02 2011Reply from BiomedTechIII
BiomedTechIII

Here at Susquehanna Health we do the same. We use the ECRI recommendation of 110 degree farenheit for blankets and 104 degree farenheit for fluids. More than that we have a policy stating these guidlines and all departments that utilize blanket warmers use this policy.



REBECCA K. CROSSLEY
BMET III | Biomedical Engineering

SUSQUEHANNA HEALTH
Report Post
Thu Nov 10 2011Reply from Clin Eng
Clin Eng

AORN covers this and the Nursing department should have a blanket warmer policy included in their regular policy manual. The temperature should be standardized throughout the institution and decided by the Nursing Administration or similar medical group. We use 130 F as the temp since lower temps do not satisfy the nurses or patients here and we perform an annual PM with this as a checkpoint on the procedure.