When turn on , the N595 do it well , but when connect the probe gives EEE6 .
Can anybody help me to repair it. ?
When turn on , the N595 do it well , but when connect the probe gives EEE6 .
Can anybody help me to repair it. ?
The service manual states that an Error 6 is "SpO2 front end reports calibration (offset) failure". I would first try a new cable then the main circuit board.
Answer provided by Timothy D., Biomed Imaging Specialist II at Rapid City Regional Hospital, via the Clinical Engineering/Healthcare Technology Management group on LinkedIN