Tue Oct 30 2012

Can anybody help me with EEE6 ?

When turn on , the N595 do it well , but when connect the probe gives EEE6 .

Can anybody  help  me  to repair it. ?


   RE: Pulse Oximeter Report Post
Thu Nov 01 2012Reply from MedWrenchManager
MedWrenchManager

The service manual states that an Error 6 is "SpO2 front end reports calibration (offset) failure". I would first try a new cable then the main circuit board.

Answer provided by Timothy D., Biomed Imaging Specialist II at Rapid City Regional Hospital, via the Clinical Engineering/Healthcare Technology Management group on LinkedIN