Tue Dec 04 2012

Clock Error code 12

According to the book Zoll states that you should replace the system board on this error. Wondering if anyone knows the really story...IE can I replace a lithium battery on the board or something. Working on equipment not subject to US hospital standards.


   RE: Defibrillator Report Post
Tue Dec 04 2012Reply from Rene66
avatar placemark

Never mind...found the answer...CR2016 system board battery. 



Rene
Report Post
Thu Aug 05 2021Reply from Kenny21
avatar placemark

Thank you for your reply! I actually need that also. 


Report Post
Thu Aug 05 2021Reply from ReNew14
avatar placemark

the coin cell on the system pcb has failed and needs to be replaced. 

99 out of 100 it's the coin cell and not the pcb. 


Report Post
Sat Aug 14 2021Reply from Dr.Mac
avatar placemark

Exactly the fix I need today- What is the easiest way to access the the coin cell?  This is a training machine I'm not interested in shipping East.


Report Post
Mon Aug 16 2021Reply from ReNew14
avatar placemark

you have to remove the 2 bottom screws under the display. this will resleae the front face. 

next you have to remove the 5 screws that hold on the upper housing and the coin cell is easily removable and visibile from that point. will need a 1632 coin cell. 

the service manual has all the instructions to remove the front face and upper housing. 