The Schiller AT-2 EKG machine at this medical office works perfectly most of the time, but will occasionally only print a flat line for a certain lead even though there is a normal connection (cable replaced, skin pad replaced, etc). Sometimes the missing lead is V3 and sometimes V6. The printhead is OK because it prints a flat line.
Thu Feb 14 2013
intermittent failure of a lead to print
Thu Feb 14 2013Reply from GaryBiomed
Did you get it happen for you on a simulator?
Here's a link to the service manual
http://photos.medwrench.com/equipmentManuals/2689-1908.pdf
Maybe you can reset to all defaults settings & do a printer alingment.
SCHILLER America Inc.
10903 NW 33rd Street
Doral, FL 33172
USA
Tel: +1 786 845 06 20
Service
www.schilleramerica.com